Library Generation for DIA Analysis

Overview
Creative Commons License: CC-BY Questions:
  • How does the peptide identification work in data independent acquisition (DIA) mass spectrometry (MS) data?

  • What is a spectral library and how to generate a spectral library using Maxquant, diapysef and OpenSwath tools?

Objectives:
  • Generating a spectral library from data dependent acquisition (DDA) MS data

  • Understanding DIA data principles and characteristics

  • Optimizing and refining a spectral library for the analysis of DIA data

Requirements:
Time estimation: 1 hour
Level: Intermediate Intermediate
Supporting Materials:
Published: Dec 2, 2020
Last modification: Nov 9, 2023
License: Tutorial Content is licensed under Creative Commons Attribution 4.0 International License. The GTN Framework is licensed under MIT
purl PURL: https://gxy.io/GTN:T00212
rating Rating: 5.0 (0 recent ratings, 2 all time)
version Revision: 10

The proteome refers to the entirety of proteins in a biological system (e.g cell, tissue, organism). Proteomics is the large-scale experimental analysis of proteins and proteomes, most often performed by mass spectrometry that enables great sensitivity and throughput. Especially for complex protein mixtures, bottom-up mass spectrometry is the standard approach. In bottom-up proteomics, proteins are digested with a specific protease into peptides and the measured peptides are in silico reassembled into the corresponding proteins. Inside the mass spectrometer, not only the peptides are measured (MS1 level), but the peptides are also fragmented into smaller peptides which are measured again (MS2 level). This is referred to as tandem-mass spectrometry (MS/MS). Identification of peptides is performed by peptide spectrum matching of the theoretical spectra generated from the input protein database (fasta file) with the measured MS2 spectra. Peptide quantification is most often performed by measuring the area under the curve of the MS1 level peptide peaks, but special techniques such as TMT and DIA allow to quantify peptides on MS2 level. Nowadays, bottom-up tandem-mass spectrometry approaches allow for the identification and quantification of several thousand proteins.

To enable the identification and quantification of such high numbers of proteins in a time-wise limited measurement there has been a lot of MS method and data acquisition optimization. The most commonly used technique is the data dependent acquisition (DDA). Here only the most abundant (top N) peptides from each MS1 scan are selected for subsequent fragmentation, with the most promising chance for successful identifications due to high intensity signals. Afterwards the respective mass over charge (m/z) values are set on an exclusion list enabling the selection of less abundant peptides from the following MS1 scan for the next cycle of top N fragmentation (MS2) scans. However, since this acquisition method is highly dependent on the abundance of the respective peptides at a given retention time this method is referred to as data dependent acquisition (DDA). This dependency on both time as well as intensity leads to measurement specific identifications between multiple measurements.

Another more recently developed method is the so called data independent acquisition (DIA). However due to some unique characteristics of this acquisition method it requieres spectral libraries, which contain the peptide sequences as well as their respective fragment spectrum (MS2) information. Here we will use DDA data from the same samples to build a spectral library which can than be used for the identification of peptides in data independent acquisition (DIA) data from the same samples using e.g. OpenSwathWorkflow OpenSwathworkflow tutorial.

DDA_lib_gen. Open image in new tab

Figure 1: Labelfree proteomics using liquid chromatography tandem-mass spectrometry (LC-MS/MS). Principles of data dependent acquisition (DDA).

High quality spectral libraries can be generated by fractionation and separate measurment of a representative sample or mixture for a cohort. During the analysis the peptide and protein identifications of all fractions can than be combined into comprehensive spectral libraries. Here the spectral libraries tend to be relativly large offering great sensitivity (even to low abundant peptides), but also contain peptides which are only observable due to the prior fractionation. Another way of generating spectral libraries is by measuring the same samples using DDA as well as using DIA. Here the DDA results will be analyzed and sample specific spectral library is generated which can than be used for the analysis of the DIA data. The resulting spectral libraries are often smaller and contain only peptides which are observable in complex and non-fractionated samples.

A plethora of software solutions were developed for the analysis of DDA proteomics data. MaxQuant is one of the most popular proteomics softwares because it is an easy to use and free software that offers functionalities for nearly all kinds of proteomics data analysis challenges Cox and Mann 2008. Mass spectrometry raw data is normally obtained in a vendor specific, proprietary file format. MaxQuant can directly take those raw files as input. For peptide identification MaxQuant uses a search engine called “Andromeda”.

The dataset in this training consists of different Spike-in mixtures of stable amounts of human peptides and different amounts of Ecoli peptides. Each sample reflects a non-fractionated MS measurement (DDA) of different Spike-in ratios. To faciliate the analysis and directly combine the identifications of all measurements, we will define the different measurements as fractions of one sample, as it would be the case for e.g. high pH, SCX or HILIC fractionated samples.

Agenda

In this tutorial, we will cover:

  1. Get data
  2. MaxQuant Analysis of DDA data
    1. Quality control results
    2. Filter for unique peptides in the evidence and msms output
  3. diapysef library generation
  4. Spectral library refinement with OpenSwathAssayGenerator
  5. Adding decoy sequences with OpenSwathDecoyGenerator
  6. Converting the spectral library with TargetedFileConverter
  7. Conclusion

Get data

Hands-on: Data upload
  1. Create a new history for this tutorial and give it a meaningful name

    To create a new history simply click the new-history icon at the top of the history panel:

    UI for creating new history

  2. Import the fasta and raw files as well as the experimental annotation and the iRTassays file from Zenodo
    https://zenodo.org/record/4293493/files/Human_database_plus_iRT.fasta
    https://zenodo.org/record/4293493/files/Ecoli_database.fasta
    https://zenodo.org/record/4293493/files/iRTassays.tsv
    https://zenodo.org/record/4293493/files/HEK_Ecoli_exp_design.txt
    https://zenodo.org/record/4293493/files/Sample1.raw
    https://zenodo.org/record/4293493/files/Sample2.raw
    https://zenodo.org/record/4293493/files/Sample3.raw
    https://zenodo.org/record/4293493/files/Sample4.raw
    https://zenodo.org/record/4293493/files/Sample5.raw
    
    • Copy the link location
    • Click galaxy-upload Upload Data at the top of the tool panel

    • Select galaxy-wf-edit Paste/Fetch Data
    • Paste the link(s) into the text field

    • Press Start

    • Close the window

  3. Once the files are green, rename the experimental design file in ‘HEK_Ecoli_exp_design’, the iRT transition file in ‘iRTassays’ and all of the the raw files in ‘Sample1.raw’, ‘Sample2.raw’, ‘Sample3.raw’, ‘Sample4.raw’ and ‘Sample5.raw’

    • Click on the galaxy-pencil pencil icon for the dataset to edit its attributes
    • In the central panel, change the Name field
    • Click the Save button

  4. Once the files are green, generate a collection for all .raw files (and name it DDA_data) as well as for both .fasta files (and name it FASTA)

    • Click on galaxy-selector Select Items at the top of the history panel Select Items button
    • Check all the datasets in your history you would like to include
    • Click n of N selected and choose Build Dataset List

      build list collection menu item

    • Enter a name for your collection
    • Click Create collection to build your collection
    • Click on the checkmark icon at the top of your history again

MaxQuant Analysis of DDA data

The MaxQuant Galaxy implementation contains the most important MaxQuant parameters. As an alternative, MaxQuant (using mqpar.xml) tool can be used with a preconfigured mqpar.xml file.

Hands-on: MaxQuant Analysis
  1. MaxQuant ( Galaxy version 1.6.10.43+galaxy3) with the following parameters:
    • In “Input Options”:
      • param-collection “FASTA files”: FASTA collection
      • “identifier parse rule”: >([^ ]*)
      • “description parse rule”: ^>.*\|.*\|[^ ]+ (.*) OS.*$
    • In “Search Options”:
      • param-file “Specify an experimental design template”: HEK_Ecoli_exp_design.txt
      • “minimum unique peptides”: 1
      • “Match between runs”: Yes
    • In “Protein quantification”:
      • “Use only unmodified peptides”: No
    • In “Parameter Group”:
      • param-collection “Infiles”: DDA_data collection
        • “missed cleavages”: 1
        • “variable modifications”: Oxidation (M)
    • “Generate PTXQC (proteomics quality control pipeline) report? (experimental setting)”: Yes
    • In “Output Options”:
      • “Select the desired outputs.”: Protein Groups Peptides mqpar.xml Evidence MSMS
    Comment: Protein Groups

    Proteins that share all their peptides with other proteins cannot be unambiguously identified. Therefore, MaxQuant groups such proteins into one protein group and only one common quantification will be calculated. The different protein properties are separated by semicolon.

Here we used a variation of the “identifier parse rule” to allow for the organism label to be present in the spectral library. For investigations of single organisms and to keep only the Uniprot identifier one can adjust the “identifier parse rule” accordingly More details on the different MaxQuant parameters can be found in this MaxQuant tutorial.

In case the MaxQuant run is not yet finished, the results can be downloaded from Zenodo to be able to continue the tutorial

  1. Import the files from Zenodo
    https://zenodo.org/record/4293493/files/MaxQuant_PTXQC_report.pdf
    https://zenodo.org/record/4293493/files/MaxQuant_Protein_Groups.tabular
    https://zenodo.org/record/4293493/files/MaxQuant_Peptides.tabular
    https://zenodo.org/record/4293493/files/MaxQuant_Evidence.tabular
    https://zenodo.org/record/4293493/files/MaxQuant_MSMS.tabular
    https://zenodo.org/record/4293493/files/MaxQuant_mqpar.xml
    

Quality control results

To get a first overview of the MaxQuant results, the PTXQC report is helpful. Click on the galaxy-eye eye of the PTXQC pdf file to open it in Galaxy. Screening through the different plots might already give you a hint on how many peptides and proteins were identified as well as some of the data quality.

The PTXQC software (Bielow et al. 2015) was built to enable direct proteomcs quality control from MaxQuant result files. This quality control can be directly used in the Galaxy MaxQuant wrapper by setting “Generate PTXQC” to yes. This will generate a pdf file with multiple quality control plots. Be aware that the cutoffs set in PTXQC might not be applicable to your experiment and mass spectrometer type and therefore “under performing” and “fail” do not necessarily mean that the quality is poor.

Question
  1. How many proteins were found in total?
  2. How many peptides were found in total?
  1. Approximately 5,370 proteins were found (see ProteinGroups output)
  2. Approximately 34,970 peptides were found (see Peptides output)

Filter for unique peptides in the evidence and msms output

Hands-on: Filtering the MaxQuant search results for unique peptides
  1. Filter with the following parameters: - param-file “Infile”: MaxQuant_Evidence
    • “With following condition”: len(c9.split(';')) < 2
    • “Number of header lines to skip”: 1
  2. Filter with the following parameters: - param-file “Infile”: MaxQuant_MSMS
    • “With following condition”: len(c12.split(';')) < 2
    • “Number of header lines to skip”: 1
    Comment: Unique Features / peptides

    Peptides which only occur in one protein are considered as unique for this specific protein. To avoid ambigious protein mapping later we filter the MaxQuant search results for unique peptides only. CAUTION: This increase in specificity (each peptide originates from only one protein) will reduce the size of the spectral library and might lead to decreased sensitivity during the DIA analysis.

Question
  1. How many unique features per lines remain after the filtering in the evidence and the msms output?
  1. ~100,000 lines in the evidence and ~98,000 lines in the msms.

diapysef library generation

Hands-on: Generation of a spectral library using the unique MaxQuant search results and indexed retention time iRT peptides
  1. diapysef library generation ( Galaxy version 0.3.5.0) with the following parameters:
    • In “MaxQuant output file msms.txt:
      • param-file “Infile”: Filter on MaxQuant_MSMS
    • In “MaxQuant output file evidence.txt:
      • param-file “Infile”: Filter on MaxQuant_Evidence
    • In “Retention time alignment method”:
      • param-file “Infile”: iRTassays.tsv
    Comment: Indexed retention time iRT peptides

    To allow for improved alignement of the different measurements, synthetic (non-endogenous) peptides which were spiked-in to all samples prior to the MS measurement. First the measured retention times (RTs) of those spiked-in peptides are extracted in the MaxQuant search results. Using the measured RTs as well as arbitrary assigned values (ranging from -26 to 99) a linear regression through the 11 synthetic peptides is applied. Based on this linear regression fit indexed retention times (iRTs) are assigned to all other peptides.

Question
  1. How many of the 11 synthetic peptides were found in the first DDA (Sample1) file?
  2. How many were found in the second DDA (Sample2) file?
  1. 8 iRT peptides were found in the first DDA file
  2. 9 iRT peptides were found in the second DDA file

Spectral library refinement with OpenSwathAssayGenerator

Hands-on: Spectral library optimization and refinement using OpenSwathAssayGenerator
  1. OpenSwathAssayGenerator ( Galaxy version 2.6+galaxy0) with the following parameters:
    • “Output file type – default: determined from file extension or content”: tabular (tsv)
    • “MZ threshold in Thomson for precursor ion selection”: 0.015
    • “upper MZ limit for precursor ions”: 1000.0
    • “MZ threshold in Thomson for fragment ion annotation”: 0.015
    Comment: Adjustment of the Spectral library

    The spectral library coming from diapysef library generation contains all observed fragment ions from the DDA runs resulting in a relativly large spectral library. Large libraries can lead to inceased processing times as well as fewer identifications after adjusting based on the False Discovery Rate (FDR). Thus, it is recommened to optimize and refine the spectral library by e.g. filtering for peptides with at least 6 transitions (increased confidence) and limiting the maximum also to 6 transitions (avoiding inflated libraries). Furthermore, the scan range can be adjusted (here between 400 - 1000 m/z) covering the same m/z range as in the DIA measurements.

Question
  1. How many transitions does the refined spectral library contain and how many were removed?
  1. The refined library contains ~170,000 lines, and almost 700,000 lines were removed.

Adding decoy sequences with OpenSwathDecoyGenerator

Hands-on: Adding decoy transitions to the spectral library
  1. OpenSwathDecoyGenerator ( Galaxy version 2.6+galaxy0) with the following parameters:
    • “Output file type – default: determined from file extension or content”: tabular (tsv)
    • “Advanced Options”: Show Advanced Options
      • “MZ threshold in Thomson for fragment ion annotation”: 0.015
    Comment: Decoy transitions

    To enable correct false discovery rate (FDR) computation later on, we add computanionally generated decoy sequences to the spectral library. Those “non-observed” sequences can be generated based on the observed sequences with slight modifications. The most commonly used methods are either shuffle (randomly altering the amino acid sequence of each observed transition) or reverse (by reversing the actually obeserved transitions). Those artificially generated transitions were labelled as decoy and are later on considered as known false positives. Example: By applying an FDR of 1 % we only allow for e.g only 1 such decoy transition out of 100 identifications. Thus one could estimate that the remaining 99 non-labeled identifications contain also 1 % false positive hits.

Question
  1. How many lines do you expect after running the OpenSwathDecoyGenerator on your spectral library containing only observed transitions?
  1. We expect double the amount of lines (~340,000), since we generate one decoy (non-observed) per target (observed) sequence.

Converting the spectral library with TargetedFileConverter

Hands-on: Converting the final spectral library from .tsv to the sqlite .pqp format
  1. TargetedFileConverter ( Galaxy version 2.6+galaxy0) with the following parameters:
    • “Output file type – default: determined from file extension or content”: pqp
    Comment: Finding the right format for the spectral library.

    Generally there is a broad variety of data formats for spectral libraries including .tsv, .dlib, .pqp and more. Depending on the DIA analysis software one or multiple formats are supported and can be used. In OpenSwathWorkflow one needs to have the spectral library in .pqp format to being able to combine multiple runs after the DIA analysis and before applying the FDR scoring. This is particularily interesting when the multiple runs should be compared later on.

Question
  1. How many peptides and proteins does the final library contain?
  2. How many target peptides and proteins are in the library?
  1. The final spectral library contains over 56,000 peptides covering over 10,200 proteins.
  2. Since we added equal numbers of decoy sequences we expect to have only half of the peptides and proteins (~28,000 and ~5,100) which are real targets.

Conclusion

DDA_lib_gen_pipe. Open image in new tab

Figure 2: All-in one workflow for DIA analysis in Galaxy. The generation of a spectral library is highlighted in red.

Using data dependent acquisiton (DDA) mass spectrometry data we generated a spectral library which can be used for the analysis of data independent acquisition (DIA) data. The quality and content of the spectral library is crucial for the following DIA data analysis. Thus, in some analysis it might be beneficial or even necessary to use slightly adjusted spectral libraries, containing e.g. semi-tryptic peptides (see enzyme in MaxQuant), non-unique peptides (skip filtering) or longer peptides (higher missed cleavages).

DDA_lib_gen_galaxy. Open image in new tab

Figure 3: Galaxy worfklow containing all necessary steps for the genration of a spectral library.